[sane-devel] Microtek ScanMaker E6 and MacOS X doesnt work
Stefan Roper
roeper at softwork-edv.de
Fri Feb 27 17:13:27 GMT 2004
Hi,
thank you for your very fast reply!
I have applied the sane-scsibuffersize.patch (more memory seldom hurts), but
with no difference.
This is my log:
goliath:~ sroeper$ export SANE_DEBUG_MICROTEK=255
goliath:~ sroeper$ scanimage -L
[sanei_debug] Setting debug level of microtek to 255.
[microtek] sane_init: MICROTEK says hello! (v0.13.1)
[microtek] sane_init: config-> # Uncomment following line to disable "real
calibration" routines...
[microtek] sane_init: config-> #norealcal
[microtek] sane_init: config-> # Uncomment following line to disable
"clever precalibration" routines...
[microtek] sane_init: config-> #noprecal
[microtek] sane_init: config-> # Using "norealcal" will revert backend to
pre-0.11.0 calibration code.
[microtek] sane_init: config-> scsi * * Scanner
[microtek] attach_scanner: iokitscsi@<01582900000000121287b191>
[microtek] attach_scanner: opening iokitscsi@<01582900000000121287b191>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<0158270000000012131e6141>
[microtek] attach_scanner: opening iokitscsi@<0158270000000012131e6141>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<01588f000000001213b2e567>
[microtek] attach_scanner: opening iokitscsi@<01588f000000001213b2e567>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<01588d000000001214403c66>
[microtek] attach_scanner: opening iokitscsi@<01588d000000001214403c66>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<015885000000001214ced725>
[microtek] attach_scanner: opening iokitscsi@<015885000000001214ced725>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<015883000000001215646d11>
[microtek] attach_scanner: opening iokitscsi@<015883000000001215646d11>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<0158fb000000001215f2e73d>
[microtek] attach_scanner: opening iokitscsi@<0158fb000000001215f2e73d>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<0158f900000000122670cd53>
[microtek] attach_scanner: opening iokitscsi@<0158f900000000122670cd53>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] sane_init: config-> /dev/scanner
[microtek] attach_scanner: /dev/scanner
[microtek] attach_scanner: opening /dev/scanner
[microtek] attach_scanner: open failed
[microtek] sane_get_devices
device `microtek:iokitscsi@<0158f900000000122670cd53>' is a Microtek
ScanMaker E6 flatbed scanner
device `microtek:iokitscsi@<0158fb000000001215f2e73d>' is a Microtek
ScanMaker E6 flatbed scanner
device `microtek:iokitscsi@<015883000000001215646d11>' is a Microtek
ScanMaker E6 flatbed scanner
device `microtek:iokitscsi@<015885000000001214ced725>' is a Microtek
ScanMaker E6 flatbed scanner
device `microtek:iokitscsi@<01588d000000001214403c66>' is a Microtek
ScanMaker E6 flatbed scanner
device `microtek:iokitscsi@<01588f000000001213b2e567>' is a Microtek
ScanMaker E6 flatbed scanner
device `microtek:iokitscsi@<0158270000000012131e6141>' is a Microtek
ScanMaker E6 flatbed scanner
device `microtek:iokitscsi@<01582900000000121287b191>' is a Microtek
ScanMaker E6 flatbed scanner
[microtek] sane_exit...
[microtek] sane_exit: MICROTEK says goodbye.
goliath:~ sroeper$ scanimage > test.pnm
[sanei_debug] Setting debug level of microtek to 255.
[microtek] sane_init: MICROTEK says hello! (v0.13.1)
[microtek] sane_init: config-> # Uncomment following line to disable "real
calibration" routines...
[microtek] sane_init: config-> #norealcal
[microtek] sane_init: config-> # Uncomment following line to disable
"clever precalibration" routines...
[microtek] sane_init: config-> #noprecal
[microtek] sane_init: config-> # Using "norealcal" will revert backend to
pre-0.11.0 calibration code.
[microtek] sane_init: config-> scsi * * Scanner
[microtek] attach_scanner: iokitscsi@<01582900000000121287b191>
[microtek] attach_scanner: opening iokitscsi@<01582900000000121287b191>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<0158270000000012131e6141>
[microtek] attach_scanner: opening iokitscsi@<0158270000000012131e6141>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<01588f000000001213b2e567>
[microtek] attach_scanner: opening iokitscsi@<01588f000000001213b2e567>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<01588d000000001214403c66>
[microtek] attach_scanner: opening iokitscsi@<01588d000000001214403c66>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<015885000000001214ced725>
[microtek] attach_scanner: opening iokitscsi@<015885000000001214ced725>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<015883000000001215646d11>
[microtek] attach_scanner: opening iokitscsi@<015883000000001215646d11>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<0158fb000000001215f2e73d>
[microtek] attach_scanner: opening iokitscsi@<0158fb000000001215f2e73d>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] attach_scanner: iokitscsi@<0158f900000000122670cd53>
[microtek] attach_scanner: opening iokitscsi@<0158f900000000122670cd53>
[microtek] attach_scanner: sending INQUIRY
[microtek] id_microtek...
[microtek] parse_inquiry...
[microtek] parse_inquiry: E6 falsely denies expanded resolution.
[microtek] parse_inquiry: E6 falsely denies 1024-byte LUT.
[microtek] dump_inquiry...
[microtek] === SANE/Microtek backend v0.13.1 ===
[microtek] ========== Scanner Inquiry Block ========mm
[microtek] 06 31 13 01 5b 16 00 00 20 20 20 20 20 20 20 20
[microtek] 53 63 61 6e 6e 65 72 20 36 30 30 20 20 20 20 20
[microtek] 31 2e 39 31 53 43 53 49 20 46 2f 57 56 33 2e 34
[microtek] 20 43 54 4c 35 33 38 30 03 4f 8c c5 03 ee 66 17
[microtek] 00 03 01 00 00 03 00 01 0e 0e 01 05 00 00 00 00
[microtek] 00 00 00 00 00 00 00 00 00 00 00 00 00 00 00 ff
[microtek] ========== Scanner Inquiry Report ==========
[microtek] ===== Scanner ID...
[microtek] Device Type Code: 0x06
[microtek] Model Code: 0x66
[microtek] Vendor Name: ' ' Model Name: 'Scanner 600 '
[microtek] Vendor Specific String: 'SCSI F/WV3.4 CTL5380'
[microtek] Firmware Rev: '1.91'
[microtek] SCSI F/W version: 3.1 Scanner F/W version: 1.3
[microtek] Response data format: 0x01
[microtek] ===== Imaging Capabilities...
[microtek] Modes: Lineart Halftone Gray Color (OnePass)
[microtek] Resolution Step Sizes: 1% 5% Expanded Resolution Support?
yes(but says no)
[microtek] Supported Bits Per Sample: 4 8
[microtek] Max. document size code: 0x03
[microtek] Max. document size: 5099 x 7799 pixels
[microtek] Frame units: pixels 1/8"'s
[microtek] # of built-in halftones: 12 Downloadable patterns? Yes
[microtek] Data Compression:
[microtek] Contrast Settings: 14 Exposure Settings: 14
[microtek] Adjustable Shadow/Highlight? yes Adjustable Midtone? yes
[microtek] Digital brightness/offset? yes
[microtek] Gamma Table Size: 1024 entries of 1 bytes (max. depth: 8)
[microtek] ===== Source Options...
[microtek] Feed type: flatbed ADF support? yes
[microtek] Document Feeder Support? yes Feeder Backtracking? yes
[microtek] Feeder Installed? yes Feeder Ready? no
[microtek] Transparency Adapter Installed? no
[microtek] Fast Color Prescan? no
[microtek] Selectable Transfer Format? no
[microtek] Color Transfer Sequence: line-by-line, non-sequential with
headers
[microtek] Three pass scan support? yes
[microtek] ModeSelect-1 and ModeSense-1 Support? yes
[microtek] Can Disable Linearization Table? no
[microtek] Can Disable Start-of-Scan Recalibration? yes
[microtek] Internal expanded expansion? no
[microtek] ====== End of Scanner Inquiry Report =======
[microtek] attach_scanner: happy.
[microtek] sane_init: config-> /dev/scanner
[microtek] attach_scanner: /dev/scanner
[microtek] attach_scanner: opening /dev/scanner
[microtek] attach_scanner: open failed
[microtek] sane_get_devices
[microtek] sane_open
[microtek] sane_open: find device...
[microtek] sane_open: create scanner...
[microtek] sane_open: initialize scanner dependent stuff...
[microtek] sane_open: Real calibration enabled.
[microtek] sane_open: init gamma tables...
[microtek] sane_open: init pass-time variables...
[microtek] sane_open: init clever cache...
[microtek] sane_open: initialize options:
[microtek] init_options...
[microtek] init_options: quant yes
[microtek] init_options: done.
[microtek] sane_control_option (opt=0,act=0,val=0xbffffbec,info=0x0)
[microtek] sane_control_option (opt=0,act=0,val=0xbffff740,info=0x0)
[microtek] sane_get_option_descriptor (0)...
[microtek] sane_get_option_descriptor (1)...
[microtek] sane_get_option_descriptor (2)...
[microtek] sane_get_option_descriptor (3)...
[microtek] sane_get_option_descriptor (4)...
[microtek] sane_get_option_descriptor (5)...
[microtek] sane_get_option_descriptor (6)...
[microtek] sane_get_option_descriptor (7)...
[microtek] sane_get_option_descriptor (8)...
[microtek] sane_get_option_descriptor (9)...
[microtek] sane_get_option_descriptor (10)...
[microtek] sane_get_option_descriptor (11)...
[microtek] sane_get_option_descriptor (12)...
[microtek] sane_get_option_descriptor (13)...
[microtek] sane_get_option_descriptor (14)...
[microtek] sane_control_option (opt=14,act=0,val=0xb34c,info=0x0)
[microtek] sane_get_option_descriptor (15)...
[microtek] sane_control_option (opt=15,act=0,val=0xb350,info=0x0)
[microtek] sane_get_option_descriptor (16)...
[microtek] sane_get_option_descriptor (17)...
[microtek] sane_get_option_descriptor (18)...
[microtek] sane_get_option_descriptor (19)...
[microtek] sane_get_option_descriptor (20)...
[microtek] sane_get_option_descriptor (21)...
[microtek] sane_get_option_descriptor (22)...
[microtek] sane_get_option_descriptor (23)...
[microtek] sane_get_option_descriptor (24)...
[microtek] sane_get_option_descriptor (25)...
[microtek] sane_get_option_descriptor (26)...
[microtek] sane_get_option_descriptor (27)...
[microtek] sane_get_option_descriptor (28)...
[microtek] sane_get_option_descriptor (29)...
[microtek] sane_get_option_descriptor (30)...
[microtek] sane_get_option_descriptor (31)...
[microtek] sane_get_option_descriptor (32)...
[microtek] sane_get_option_descriptor (33)...
[microtek] sane_control_option (opt=12,act=0,val=0xbffff744,info=0x0)
[microtek] sane_control_option (opt=13,act=0,val=0xbffff744,info=0x0)
[microtek] sane_control_option (opt=12,act=0,val=0xbffffbf0,info=0x0)
[microtek] sane_get_option_descriptor (14)...
[microtek] sane_control_option (opt=14,act=1,val=0xbffffbf4,info=0xbffff768)
[microtek] sane_control_option (opt=13,act=0,val=0xbffffbf0,info=0x0)
[microtek] sane_get_option_descriptor (15)...
[microtek] sane_control_option (opt=15,act=1,val=0xbffffbf4,info=0xbffff768)
[microtek] sane_start...
[microtek] sane_get_parameters...
[microtek] sane_get_parameters: regular 1-pass color
[microtek] sane_get_parameters: res_code = 16 (10)
[microtek] bright_r of 0 set to 0x0
[microtek] sane_get_parameters: dots_per_mm: 3.937008
[microtek] sane_get_parameters: units_per_mm: 23.622047
[microtek] WIDTHPIX: before exp: 850
[microtek] sane_get_parameters: lines: 1300 ppl: 850 bpl: 2550
[microtek] .wait_ready 0...
[microtek] finagle_precal...
[microtek] .compare_mode_sense 0...
[microtek] CMS: 0( 0) 64( 0) 8( 0) 4( 0) 7( 0) 7( 0) 0( 0) ff( 0)
80( 0) 7( 0) 7( 0) 0( 0) ff( 0) 80( 0) 7( 0) 7( 0) 0( 0) ff( 0) 80( 0)
[microtek] finagle_precal: must precalibrate!
[microtek] do_real_calibrate...
[microtek] .start_calibrate...
[microtek] STCal:11 0 0 0 2 0 1 a
[microtek] .get_scan_status 0...
[microtek] get_scan_status(0): 0, 0, 0 -> #0
[microtek] > 0 0 0 0 0 0
[microtek] do_real_cal: getting data (max=0)
[microtek] ...nleft 12 toget 0 size 0 spot 0 input+spot 0x3065a0
[microtek] .read_scan_data...
[microtek] ...nleft 12 toget 0 size 0 spot 0 input+spot 0x3065a0
[microtek] .read_scan_data...
[microtek] ...nleft 12 toget 0 size 0 spot 0 input+spot 0x3065a0
[microtek] .read_scan_data...
[microtek] ...nleft 12 toget 0 size 0 spot 0 input+spot 0x3065a0
[microtek] .read_scan_data...
[microtek] ...nleft 12 toget 0 size 0 spot 0 input+spot 0x3065a0
[microtek] .read_scan_data...
...
(will be repeated endless)
...
Seems to be a different thing...
Any further suggestions ?
Thanks,
Stefan
-----Ursprungliche Nachricht-----
Von: Mattias Ellert [mailto:mattias.ellert at tsl.uu.se]
Gesendet: Freitag, 27. Februar 2004 14:43
An: roeper at softwork-edv.de
Betreff: Re: [sane-devel] Microtek ScanMaker E6 and MacOS X doesnt work
Stefan Roper wrote:
> Hi,
>
> i tried the lastest ( Version 1.0.13) sane binary packages on my Mac G4
> with an Adaptec 2906 Scsi card and the Microtek ScanMaker E6.
>
> The suggested test scan with the command:
> $ scanimage -d test:0 > test.pnm
> works.
>
> The command:
> $ sane-find-scanner
>
> # No SCSI scanners found. If you expected something different, make sure
> that
> # you have loaded a SCSI driver for your SCSI adapter.
>
> # No USB scanners found. If you expected something different, make sure
that
> # you have loaded a driver for your USB host controller and have installed
a
> # kernel scanner module.
>
> # Scanners connected to the parallel port or other proprietary ports
> can't be
> # detected by this program.
>
> # You may want to run this program as root to find all devices. Once you
> # found the scanner devices, be sure to adjust access permissions as
> # necessary.
>
> doesn't find anything, which may be ok because the Mac has no
> /dev/scanner device.
>
> So i tried the next command:
> $ scanimage -L
> device `microtek:iokitscsi@<015909000000000f0137c50b>' is a Microtek
> ScanMaker E6 flatbed scanner
> device `microtek:iokitscsi@<01590b000000000e2c5655a3>' is a Microtek
> ScanMaker E6 flatbed scanner
> device `microtek:iokitscsi@<015893000000000e2bc9b882>' is a Microtek
> ScanMaker E6 flatbed scanner
> device `microtek:iokitscsi@<015895000000000e2b39b540>' is a Microtek
> ScanMaker E6 flatbed scanner
> device `microtek:iokitscsi@<01589d000000000e2aa2e3dc>' is a Microtek
> ScanMaker E6 flatbed scanner
> device `microtek:iokitscsi@<01589f000000000e2a0de2fe>' is a Microtek
> ScanMaker E6 flatbed scanner
> device `microtek:iokitscsi@<015837000000000e2984a549>' is a Microtek
> ScanMaker E6 flatbed scanner
> device `microtek:iokitscsi@<015839000000000e28f1ee11>' is a Microtek
> ScanMaker E6 flatbed scanner
>
> Looks like there is a scanner detected, but why 8 times ?
>
> An scanimage > test.pnm doesn't work, the scanner starts, but then the
> process hangs. After killing the process, the mac sometimes hangs.
>
> Any suggestions for me ?
>
> Thanks in advance,
> Stefan
Activate debugging of the backend:
for bash: SANE_DEBUG_MICROTEK=255 scanimage > test.pnm
for tcsh: (setenv SANE_DEBUG_MICROTEK 255; scanimage > test.pnm)
Most probable cause for failure in the microtek backend is:
[microtek] sane_start: SCSI buffer smaller that one scan line!
[microtek] end_scan...
[microtek] .stop_scan...
If this is the case for you, please try the suggested patch on
http://www.ellert.se/twain-sane/patches.html
(Whenever I get back to Uppsala I will create a new set of packages...)
Mattias
--
________________________________________________________________________
mattias.ellert at tsl.uu.se tel: +46 18 471 32 58
http://www.tsl.uu.se/~ellert/ fax: +46 18 471 35 13
________________________________________________________________________
More information about the sane-devel
mailing list