Hello,
You can find the result of this test here :
http://ggastebois.free.fr/lide90_snoop/no_shading_2bright_2dark.tar
The calibration is about 25s long.
Resulting image is bad (as befor).
Regards
Guillaume
Pierre Willenbrock a écrit :
> Pierre Willenbrock schrieb:
>> Guillaume Gastebois schrieb:
>>> Hello,
>>>
>>> You can find all of that on :
>>> http://ggastebois.free.fr/lide90_snoop/no_shading.tar
>>>
>>> What seems strange in calibration ?
>>>
>> The minimum/maximum values retrieved from the scans do not scale with
>> the offset values, but in a rather random manner. The calibration
>> expects a linear relationship.
>>
>> Regards,
>> Pierre
>>
>
> Hi,
>
> the raw offset calibration data looks horrible. Too bright and noisy.
> Please change the too bright/too dark thresholds for
> gl841_offset_calibration in genesys_gl841.c like this:
>
> @@ -4545,9 +4546,9 @@
> val =
> first_line[i * 2 * channels + 2 * j + 1] * 256 +
> first_line[i * 2 * channels + 2 * j];
> - if (val < 10)
> + if (val < 1000)
> cmin[j]++;
> - if (val > 65525)
> + if (val > 40000)
> cmax[j]++;
> }
>
> Send/Put online the debug files offset*.pnm, and the
> "gl841_offset_calibration: acceptable offsets: " lines from the debug
> output(Or the whole set, whichever is more convenient ;-)).
>
> Regards,
> Pierre
>
>